A cautionary tale that others might benefit from.
I've got an Hitachi V422 'scope and was in the habit of storing it vertically on its base, tube face / controls face up, when not in use to save bench space.
Recently I started to experience problems with intermittent contacts on the push/pull switches and pot switches.
On investigation I discovered that the shaft lubrication had migrated downwards onto the switch contacts and formed a non conducting barrier.
It was a devil's own job to re-assemble the switch contacts after cleaning, you needed 4 hands.
So the solution to the problem is only ever to store the instrument in the operational position, ie with scope face and controls facing forward.
If you are ever offered a s/h unit it's worth checking the operation of these controls.
If it's suffering from this problem they will not self-clean despite repeated operation.
Switch cleaner might work, but getting into the switch bodies will be a problem.
Cheers,
Graham